PQIA Noack test temp

It's going to change for GF-7 too due to poor correlation with oil consumption in the field.
 
Last edited:
There is a discussion of the different procedures within the ASTM test here, in particular the linked white paper for details on the temperature and how the different allowed procedures compare:

https://www.savantlab.com/testing-highlights/engine-oil-volatility-and-astm-d5800/
And in there lies the answer, in one of the links:
https://www.savantlab.com/documents/58/VolatilityWhitePaper03-Selby.pdf

Thus, operating temperature necessary at the sensor in this configuration (later called Procedure B) had to be lower than the original Woods Metal Noack value of 250.0°C. Not surprisingly, it was found by the developers of this instrument that 245.2 ± 0.5°C was needed in order to approach Woods Metal Noack results.
 